DICOM Conformance Statement FORUM
|
|
|
- Roxanne Freeman
- 10 years ago
- Views:
Transcription
1 DICOM Conformance Statement FORUM Version 3.1 Carl Zeiss Meditec AG Goeschwitzerstraße Jena Germany Document: DICOM Conformance Statement_FORUM_3.1.doc Page 1 of 25
2 1 Conformance Statement Overview This document is structured as suggested in the DICOM Standard (PS 3.2, 2011). SOP Classes User of Service (SCU) Provider of Service (SCP) Verification Verification Yes Yes Transfer Image Storage Classes Secondary Capture Image Storage Yes Yes Multi-frame Grayscale Byte Secondary Capture Image Storage Yes Yes Multi-frame True Color Secondary Capture Image Storage Yes Yes Raw Data Storage Yes Yes Visible Light Endoscopic Image Storage Yes Yes Visible Light Microscopic Image Storage Yes Yes Visible Light Photographic Image Storage Yes Yes Ophthalmic Photography 8 Bit Image Storage Yes Yes Ophthalmic Tomography Image Storage Yes Yes Transfer Video Image Storage Classes Video Endoscopic Image Storage Yes Yes Video Microscopic Image Storage Yes Yes Video Photographic Image Storage Yes Yes Transfer Other Storage Classes Encapsulated PDF Storage Yes Yes Query/Retrieve Patient Root Q/R Information Model FIND No Yes Study Root Query/Retrieve Information Model FIND No Yes Study Root Query/Retrieve Information Model MOVE No Yes Workflow Management Modality Worklist Information Model - FIND No Yes Storage Commitment Push Model No Yes FORUM does not support Media Interchange. Document: DICOM Conformance Statement_FORUM_3.1.doc Page 2 of 25
3 2 Table of Contents 1 Conformance Statement Overview Table of Contents Introduction Revision History Audience Remarks Definitions, Terms and Abbreviations References Networking Implementation Model Application Data Flow Functional Definition of AEs Verification Receive Images Storage Commitment Query from other devices Retrieve to other devices Store to other devices Modality Worklist AE Specifications FORUM Application Entity Specification SOP Classes Associations Policies General Number of Associations Asynchronous Nature Implementation Identifying Information Association Initiation Policy Activity Verification (SCU) Activity Store Objects (SCU) Association Acceptance Policy Activity Verification (SCP) Activity Store Objects (SCP) Activity Request Storage Commitment (SCP) Activity Find Objects (SCP) Activity Move Objects (SCP) Activity Query Modality Worklist (SCP) Network Interfaces Physical Network Interface Additional Protocols Configuration AE Title/Presentation Address Mapping Local AE Titles Remote AE Titles Logging Media Interchange Support Of Character Sets Accepted Character Sets Returned Character Sets Security Annexes IOD Contents Created SOP Instance(s) Usage of Attributes from Received IODs Attribute Mapping Coerced/Modified Files Patient Information Missing UIDs Patient Identification Conflict with Study Instance UID Data Dictionary of Private Attributes Private Transfer Syntaxes Document: DICOM Conformance Statement_FORUM_3.1.doc Page 3 of 25
4 3.1 Revision History 3 Introduction Document Author Date Change Revision 1.0 tku Initial revision for Forum v tku Adjustments for Forum v gws Revision for FORUM v2.5 (merged from CG-LEH-MS- 203-DICOM Conformance Statement 2.5) 2.6 gws Revision for FORUM v2.6 (merged from CG-LEH-MS- 254-DICOM Conformance Statement 2.6.doc) 3.0 gws Revision for FORUM v gws Revision for FORUM v Audience This document is intended for hospital staff, health system integrators, software designers or implementers. The reader should have a basic understanding of DICOM. 3.3 Remarks If another device matches this conformance statement based on the comparison with its own conformance statement, there is a chance, but no guarantee, that they interoperate. DICOM only deals with communication; it does not specify what is needed for certain applications to run on a device. 3.4 Definitions, Terms and Abbreviations Abbreviation AE AET DICOM ILE IOD JPG-1 MWL MPG2 NB RLE SCP SCU SOP SW TCP/IP UID VL Definition Application Entity Application Entity Title Digital Imaging and Communications in Medicine Implicit Little Endian Information Object Definition JPEG Coding Process 1 transfer syntax; JPEG Baseline; ISO Modality Work List Motion Picture Expert Group 2; Abbreviation and synonym for video encoding and compression transfer syntax. Network Broker Run Length Encoding Service Class Provider Service Class User Service Object Pair, pair of user and provider. Software Transmission Control Protocol / Internet Protocol Unique Identifier Visible Light 3.5 References NEMA PS3 / ISO 12052, Digital Imaging and Communications in Medicine (DICOM) Standard, National Electrical Manufacturers Association, Rosslyn, VA, USA (available free at Integrating the Healthcare Enterprise (IHE) EYECARE Technical Framework, rev 3.7, 2010 (available free at Document: DICOM Conformance Statement_FORUM_3.1.doc Page 4 of 25
5 4.1 Implementation Model Application Data Flow 4 Networking The FORUM application can be configured in a way that the complete functionality is covered by one AE. The diagram above shows the configuration where Storage related activities and Worklist related activities are separate AEs. Document: DICOM Conformance Statement_FORUM_3.1.doc Page 5 of 25
6 4.1.2 Functional Definition of AEs Verification Both the FORUM AE and the FORUM MWL AE will respond to a DICOM C-Echo request from a remote AE in order to verify the communication. The FORUM AE can also initiate an association and request verification to a remote AE Receive Images The FORUM AE stores a received image in its internal data storage. The FORUM AE stores each image with the File Meta Information attached to it. The FORUM AE extracts the query information with respect to the patient, study, series, and instance, and stores this information in its internal index database Storage Commitment The FORUM AE acts as an SCP for Storage Commitment to take over the responsibility explicitly for storing received DICOM instances Query from other devices The FORUM AE responds to queries based on the information stored in its index database. When a DICOM query request is received, the FORUM AE converts the request into a corresponding database query. The results of the database query are then returned to the remote AE Retrieve to other devices The FORUM AE acts as an SCP for C-Move requests to retrieve DICOM instances Store to other devices The FORUM AE acts as an SCU for C-Store requests to transmit DICOM instances to a remote DICOM AE Modality Worklist The FORUM MWL AE returns a Modality Worklist item in response to a request from a remote AE. Worklist items to be returned, can be created - Manually using the FORUM MWL Scheduler UI - Automatically through the EMR XML interface - Automatically through the HL7 interface This information is cached in the local database. When a DICOM MWL request is received, the FORUM MWL AE converts the request into a database query. The results of the database query are then returned via DICOM (as a MWL response) to the remote AE. Document: DICOM Conformance Statement_FORUM_3.1.doc Page 6 of 25
7 4.2 AE Specifications FORUM Application Entity Specification SOP Classes SOP Class Name SOP Class UID SCU SCP Verification Yes Yes Modality Worklist Information Model - FIND No Yes Ophthalmic Photography 8 Bit Image Storage Yes Yes Encapsulated PDF Storage Yes Yes Secondary Capture Image Storage Yes Yes Multi-frame Grayscale Byte Secondary Capture Yes Yes Image Storage Multi-frame True Color Secondary Capture Image Yes Yes Storage Raw Data Storage Yes Yes Visible Light Endoscopic Image Storage Yes Yes Visible Light Microscopic Image Storage Yes Yes Visible Light Photographic Image Storage Yes Yes Video Endoscopic Image Storage Yes Yes Video Microscopic Image Storage Yes Yes Video Photographic Image Storage Yes Yes Ophthalmic Tomography Image Storage Yes Yes Patient Root Q/R Information Model FIND No Yes Study Root Q/R Information Model FIND No Yes Study Root Q/R Information Model MOVE No Yes Storage Commitment Push Model No Yes Associations Policies General DICOM standard Application Context Name is DICOM 3.0. Application Context Name Number of Associations The FORUM SW can be configured to use either one AE or two separate AEs for Storage and Worklist related activities. The maximum number of association applies per AE. Maximum number of simultaneous associations 50 per AE Asynchronous Nature FORUM does not support asynchronous communication (multiple outstanding transactions over a single Association) Implementation Identifying Information Implementation Class UID Implementation Version Name FORUM 3.1 Document: DICOM Conformance Statement_FORUM_3.1.doc Page 7 of 25
8 Association Initiation Policy Activity Verification (SCU) Description and Sequencing of Activities The FORUM AE sends DICOM Verification requests when the user requests a test of validity of a DICOM connection through the GUI Proposed Presentation Contexts Presentation Context Table Abstract Syntax Transfer Syntax Role Ext. Name UID Name UID List Neg. List Verification ILE SCU No SOP Specific Conformance Verification The FORUM AE provides standard conformance to the DICOM Verification Service Class as an SCU Activity Store Objects (SCU) Description and Sequencing of Activities The FORUM AE will transmit instances either driven by user requests through the UI, by an automated forwarding mechanism or as reaction to a received C-MOVE request. The automatic forwarding can be configured by the user through the UI. An association is established when the user or the forwarding mechanism initiates a transmit request. The FORUM AE will establish an association automatically in response to a C-MOVE request Proposed Presentation Contexts The FORUM AE may request any of the Presentation Contexts listed in the following table for Storage. The FORUM AE will only propose the transfer syntax, which was used when the object was initially accepted by the server. Presentation Context Table Abstract Syntax Transfer Syntax Role Ext. Neg. Name / UID Name List UID List All SOP classes listed in section "Transfer Image Storage Classes" All SOP classes listed in section "Transfer Video Image Storage Classes" All SOP classes listed in section "Transfer Other Storage Classes" Implicit VR Little Endian Explicit VR Little Endian Explicit VR Big Endian RLE Lossless JPEG Baseline JPEG Lossless JPEG 2000 (Lossless Only) JPEG 2000 MPEG2 MPML MPEG2 MPHL MPEG2 MPML MPEG2 MPHL Implicit VR Little Endian Explicit VR Little Endian Explicit VR Big Endian SCU SCU SCU No No No SOP Specific Conformance Store Objects The FORUM AE provides limited Standard conformance to the DICOM Storage Service Class as an SCU: Neither is it guaranteed that the DICOM Default Transfer Syntax (Implicit VR Little Endian) is always proposed in a presentation context. Nor is an on-the-fly transcoding to the DICOM Default Transfer Syntax supported for instances received and stored in lossless compression. A successful C-Store response status will not generate any actions. Document: DICOM Conformance Statement_FORUM_3.1.doc Page 8 of 25
9 Association Acceptance Policy Activity Verification (SCP) Description and Sequencing of Activities Both the FORUM AE and the FORUM MWL AE will respond to Verification requests from a Verification SCU. This way the requesting SCU can determine whether the respective FORUM AEs are able to receive and handle incoming DICOM requests Accepted Presentation Contexts Presentation Context Table Abstract Syntax Transfer Syntax Role Ext. Name UID Name UID List Neg. List Verification ILE SCP No SOP Specific Conformance Verification FORUM provides standard conformance to the DICOM Verification Service Class as an SCP. One of the following status codes are returned: Service Status Further Meaning Error Code Behavior Success Success 0000 Operation performed properly. Both the FORUM AE and the FORUM MWL AE will always accept a Presentation Context for the Verification SOP Class with the default DICOM transfer syntax Activity Store Objects (SCP) Description and Sequencing of Activities The FORUM AE will store DICOM instances that are sent to it from a remote SCU. All images received by the FORUM AE can be retrieved at a later time from the FORUM AE Accepted Presentation Contexts Presentation Context Table Abstract Syntax Transfer Syntax Role Ext. Neg. Name / UID Name List UID List All SOP classes listed in section "Transfer Image Storage Classes" All SOP classes listed in section "Transfer Video Image Storage Classes" All SOP classes listed in section "Transfer Other Storage Classes" Implicit VR Little Endian Explicit VR Little Endian Explicit VR Big Endian RLE Lossless JPEG Baseline JPEG Lossless JPEG 2000 (Lossless Only) JPEG 2000 MPEG2 MPML MPEG2 MPHL MPEG2 MPML MPEG2 MPHL Implicit VR Little Endian Explicit VR Little Endian Explicit VR Big Endian SCP SCP SCP No No No Document: DICOM Conformance Statement_FORUM_3.1.doc Page 9 of 25
10 SOP Specific Conformance Store Objects The FORUM AE conforms to the DICOM Storage Service Class as a Level 2 (Full) SCP. No elements are discarded. Certain Attribute will be coerced, see Coerced/Modified Files. All Type 1, Type 2 and Type 3 attributes will be retained. Private attributes will be stored and included when the object is sent out later to a remote AE. In case an instance is rejected, one of the following status codes is returned by the FORUM AE: Service Status Further Meaning Error Code Description Refused Out Of Resources A700 Error while storage space is less than 3GB. Missing license for MPEG No Such SOP class 0118 If a SOP Class is used, that was not negotiated. Unrecognized Operation 0211 Unrecognized value in Command Filed Error Invalid Attribute Value 0106 Unsupported SOP Instance UID prefix. MpegFormatException Processing Failure 0110 Unexpected internal errors. Duplicate SOP instance 0111 SOP Instance UID is already present in this archive while a new C-STORE request is being processed. Invalid Object Instance 0117 The series of an incoming C-STORE request is already attached to another study. Missing Attribute Value 0121 An Instance UID is missing. SOP class Not Supported 0122 Incoming SOP Class is not supported by FORUM Unable To Process C000 Unexpected internal errors. Success Success 0000 Operation performed properly. The FORUM AE will accept any number of Presentation Contexts per association request. Any single Abstract Syntax may be specified more than once in an association request, if the Transfer Syntaxes differ between the Presentation Contexts. Explicit VR Little Endian is preferred over Implicit VR Little Endian as the accepted Transfer Syntax when both Transfer Syntaxes are offered in a Presentation Context in an Associate Request. FORUM AE will only accept association requests for known (configured) remote AEs Activity Request Storage Commitment (SCP) Description and Sequencing of Activities The FORUM AE stores instances that are sent to it from a remote SCU. The request for Storage Commitment may then be transmitted to the FORUM AE together with a list of references to one or more DICOM SOP instances Accepted Presentation Contexts Presentation Context Table Abstract Syntax Transfer Syntax Role Ext. Name UID Name UID List Neg. List Storage Commitment Push Model ILE SCP No Document: DICOM Conformance Statement_FORUM_3.1.doc Page 10 of 25
11 SOP Specific Conformance Storage Commitment The FORUM AE provides conformance to the DICOM Storage Commitment Service Class as an SCP. The number of referenced instances within one request is limited to 500. The N-EVENT-REPORT message is always sent in a separate association from the N-ACTION operation. For Storage Commitment operations, the FORUM AE will only accept association requests for known (configured) remote AEs. One of the following Status codes is returned by the FORUM AE: Service Status Further Meaning Error Code Description Error Processing Failure 0110 Unexpected internal errors or too many items in the Referenced SOP Sequence (at least more than 500) Success Success 0000 Operation performed properly. The N-EVENT-REPORT contains the following Failure Reasons in the Failed SOP Sequence: Failure Reason Further Meaning Error Code Description Error No such object instance 0112 The SOP instance is not available in FORUM. Class / Instance conflict 0119 The SOP instance is stored with another SOP class UID in FORUM Referenced SOP Class not supported 0122 The SOP instance is not available in FORUM and the SOP Class is not supported by FORUM. Document: DICOM Conformance Statement_FORUM_3.1.doc Page 11 of 25
12 Activity Find Objects (SCP) Description and Sequencing of Activities The FORUM AE will respond to query requests that are sent to it by a remote SCU Accepted Presentation Contexts Presentation Context Table Abstract Syntax Transfer Syntax Role Ext. Name UID Name UID List Neg. List Patient Root Q/R ILE SCP Yes Information Model FIND Study Root Q/R Information Model FIND ILE SCP Yes C-Find Extended Negotiation is supported: For the fields "Relational-queries" and "Date-time matching", the value 1 is returned, meaning that those are supported by the FORUM AE SOP Specific Conformance Find Objects The FORUM AE provides standard conformance to the DICOM Query/Retrieve Service Class as an SCP. The FORUM AE supports both Hierarchical Queries and the Relational-queries extended SCP behavior. Matching for PN VR attributes is always done case-insensitive. The following tables contain detailed information on matching and returned keys. Patient Root / Patient Level Attributes: Description Tag Matching Type Patient's Name (0010,0010) U S * Patient ID (0010,0020) U S *, UNIQUE together with Issuer of Patient ID Issuer of Patient ID (0010,0021) U S *, UNIQUE together with Patient ID Other Patient IDs (0010,1000) U S * Patient's Birth Date (0010,0030) U S R Patient's Sex (0010,0040) U S Ethnic Group (0010,2160) U S Patient Comments (0010,4000) U S * Patient Root / Study Level Attributes: Description Tag Matching Type Study Instance UID (0020,000D) U S, UNIQUE Study ID (0020,0010) U S Accession Number (0008,0050) U S * Study Date (0008,0020) U CDT Study Time (0008,0030) U CDT Modalities in Study (0008,0061) U S Referring Physician s Name (0008,0090) U S * Study Description (0008,1030) U S * Admitting Diagnoses Description (0008,1080) U S * Document: DICOM Conformance Statement_FORUM_3.1.doc Page 12 of 25
13 Study Root / Study Level Attributes: Description Tag Matching Type Study Instance UID (0020,000D) U S, UNIQUE Study ID (0020,0010) U S Accession Number (0008,0050) U S * Study Date (0008,0020) U CDT Study Time (0008,0030) U CDT Modalities in Study (0008,0061) U S Referring Physician s Name (0008,0090) U S * Study Description (0008,1030) U S * Admitting Diagnoses Description (0008,1080) U S * Patient's Name (0010,0010) U S * Patient ID (0010,0020) U S * Issuer of Patient ID (0010,0021) U S * Other Patient IDs (0010,1000) U S * Patient's Birth Date (0010,0030) U S R Patient's Sex (0010,0040) U S Ethnic Group (0010,2160) U S Patient Comments (0010,4000) U S * Series Level Attributes: Description Tag Matching Type Modality (0008,0060) U S Series Number (0020,0011) U S Series Instance UID (0020,000E) U S, UNIQUE Series Date (0008,0021) U CDT Series Time (0008,0031) U CDT Series Description (0008,103E) U S * Laterality (0020,0060) U S Number of Series Related Instances (0020,1209) return key only Performing Physicians Name (0008,1050) U S * Manufacturer's Model Name (0008,1090) U S * Performed Procedure Step Start Date (0040,0244) U CDT Performed Procedure Step Start Time (0040,0245) U CDT Request Attributes Sequence (0040,0275) Requested Procedure ID (0040,1001) U S Scheduled Procedure Step ID (0040,0009) U S Document: DICOM Conformance Statement_FORUM_3.1.doc Page 13 of 25
14 Instance Level Attributes: Description Tag Matching Type Image Type (0008,0008) U S * Instance Number (0020,0013) U S SOP Instance UID (0008,0018) U S, UNIQUE SOP Class UID (0008,0016) U S Instance Creation Date (0008,0012) U CDT Instance Creation Time (0008,0013) U CDT Acquisition Date Time (**) (0008,002A) U CDT Icon Image Sequence (0088,0200) return key only Image Laterality (0020,0062) U S Document Title (0042,0010) U S * Referenced Image Sequence (*) (0008,1140) Referenced SOP Instance UID (0008,1155) U S Referenced SOP Class UID (0008,1150) return key only Purpose Of Reference Code Sequence (0040,A170) Coding Scheme Designator (0008,0102) U S Coding Scheme Version (0008,0103) U S Code Value (0008,0100) U S Referenced Instance Sequence (*) (0008,114A) Referenced SOP Instance UID (0008,1155) U S Referenced SOP Class UID (0008,1150) return key only Purpose Of Reference Code Sequence (0040,A170) Coding Scheme Designator (0008,0102) U S Coding Scheme Version (0008,0103) U S Code Value (0008,0100) U S Source Instance Sequence (*) (0042,0013) U S Referenced SOP Instance UID (0008,1155) U S Referenced SOP Class UID (0008,1150) return key only Purpose Of Reference Code Sequence (0040,A170) Coding Scheme Designator (0008,0102) U S Coding Scheme Version (0008,0103) U S Code Value (0008,0100) U S Remark (*): Only one of "Referenced Image Sequence", "Referenced Instance Sequence" or " Source Instance Sequence" is supported in one C-FIND_REQUEST Remark (**): The responses are sorted descending by Acquisition Date Time. Matching Types: U Universal Matching S Single value Matching * - Wildcard Matching R Range Matching CDT Combined Date Time Matching UNIQUE Unique query key for that level Document: DICOM Conformance Statement_FORUM_3.1.doc Page 14 of 25
15 The FORUM AE returns one of the following status codes in a C-FIND response: Service Status Further Meaning Error Code Description Refused No Such SOP class 0118 If a SOP Class is used, that was not negotiated. Unrecognized Operation 0211 Unrecognized value in Command Filed Failed Invalid Attribute Value 0106 Wrong query retrieve level Processing Failure 0110 Unexpected internal errors. Unable To Process C000 Unexpected internal errors. Cancel FE00 The original requester canceled this operation. Pending Pending FF00 All Optional Keys are supported in the same manner as Required Keys. Warning Too Many Results C001 If the received request leads to more than 5000 responses then these 5000 responses will be returned followed by this status. On Query Retrieve Level "IMAGE" the most recent SOP instances according to Acquisition Date Time will be returned. Success Success 0000 Operation performed properly. For C-FIND operations, the FORUM AE will only accept association requests for known (configured) remote AEs. FORUM does not support matching on separate components for Attributes with a PN Value Representation, and matches always and only the first component. A conflicted patient in FORUM will be generated, if an incoming C-STORE dataset: - contains no patient id. - contains an existing combination of issuer of patient id and a patient id and this existing patient is not a "corresponding patient" as defined in " Patient Information". A Patient Root Query in FORUM returns only unconflicted patients. A Study Root Query in FORUM returns all patients. Document: DICOM Conformance Statement_FORUM_3.1.doc Page 15 of 25
16 Activity Move Objects (SCP) Description and Sequencing of Activities The FORUM AE will respond to C-MOVE requests that are sent to it by a remote SCU. The FORUM AE will establish a new Association with the remote AE specified in the Move Destination for the C- STORE sub operation. The FORUM AE will always only propose the transfer syntax used when the object was initially sent to the server Accepted Presentation Contexts Presentation Context Table Abstract Syntax Transfer Syntax Role Ext. Name UID Name UID List Neg. List Study Root Q/R Information Model MOVE ILE SCP No SOP Specific Conformance Move Objects The FORUM AE will establish an association with the destination specified in the C-MOVE request. One or more of the Presentation Contexts listed in chapter may be negotiated in this association. The following tables contain detailed information on supported matching keys: Study Root / Study Level Attributes: Description Tag Matching Type Study Instance UID (0020,000D) List of UID Matching Series Level Attributes: Description Tag Matching Type Series Instance UID (0020,000E) List of UID Matching Instance Level Attributes: Description Tag Matching Type SOP Instance UID (0008,0018) List of UID Matching The Query/Retrieve Level is identified implicit by the used key. If a non-empty SOP Instance UID is present, then those will be retrieved, else if a non-empty Series Instance UID is present, then those will be retrieved, else if a non-empty Study Instance UID is present, then those will be retrieved. The FORUM AE returns one of the following status codes in a C-MOVE response: Service Status Further Meaning Error Code Description Refused No Such SOP class 0118 If a SOP Class is used, that was not negotiated. Unrecognized Operation 0211 Unrecognized value in Command Filed Error Processing Failure 0110 Unexpected internal errors. No Such Object Instance 0112 C-MOVE request for non existing SOP instance Unable To Process C000 Unexpected internal errors. Success Success 0000 Operation performed properly. Document: DICOM Conformance Statement_FORUM_3.1.doc Page 16 of 25
17 Activity Query Modality Worklist (SCP) Description and Sequencing of Activities The FORUM MWL AE will respond to DICOM C-FIND requests in response to a remote AE querying for a Modality Worklist Accepted Presentation Contexts Presentation Context Table Abstract Syntax Transfer Syntax Role Ext. Name UID Name UID List Neg. List Modality Worklist Information Model - FIND ILE SCP Yes C-Find Extended Negotiation is supported: For the field "Date-time matching" the value 1 is returned, meaning that this is supported by the FORUM AE SOP Specific Conformance MWL The FORUM MWL AE provides standard conformance to the DICOM Basic Worklist Management Service Class. The FORUM application can be configured to use only one AE for MWL and all the other described transactions. The following table contains detailed information on matching and returned keys: Description Tag Matching Type Accession Number (0008,0050) U S Requested Procedure ID (0040,1001) U S Scheduled Procedure Step Sequence (0040,0100) Scheduled Station AE Title (0040,0001) U S Scheduled Procedure Step Start Date (0040,0002) U CDT Scheduled Procedure Step Start Time (0040,0003) U CDT Modality (0008,0060) U S Patient's Name (0010,0010) U S * Patient ID (0010,0020) U S * Issuer of Patient ID (0010,0021) U S * Patient's Birth Date (0010,0030) U S R Patient's Sex (0010,0040) U S Patient Comments (0010,4000) U S * Document: DICOM Conformance Statement_FORUM_3.1.doc Page 17 of 25
18 The following table contains the supported return keys: Description Tag Scheduled Procedure Step Sequence (0040,0100) >Scheduled Station AE Title (0040,0001) >Scheduled Procedure Step Start Date (0040,0002) >Scheduled Procedure Step Start Time (0040,0003) >Modality (0008,0060) >Scheduled Performing Physician's Name (0040,0006) >Scheduled Procedure Step Description (0040,0007) > Scheduled Protocol Code Sequence (0040,0008) >> Code Meaning (0008,0104) >> Code Value (0008,0100) >> Coding Scheme Designator (0008,0102) >> Coding Scheme Version (0008,0103) >Scheduled Procedure Step ID (0040,0009) Requested Procedure ID (0040,1001) Requested Procedure Comments (0040,1400) Requested Procedure Description (0032,1060) Study Instance UID (0020,000D) Study Date (=Scheduled Procedure Step Start Date) (0008,0020) Study Time (=Scheduled Procedure Step Start Time) (0008,0030) Accession Number (0008,0050) Requesting Physician (0032,1032) Requested Procedure Code Sequence (=Scheduled Protocol (0032,1064) Code Sequence) > Code Meaning (0008,0104) > Code Value (0008,0100) > Coding Scheme Designator (0008,0102) > Coding Scheme Version (0008,0103) Referring Physician s Name (0008,0090) Patient's Name (0010,0010) Patient ID (0010,0020) Issuer of Patient ID (0010,0021) Patient's Birth Date (0010,0030) Patient's Sex (0010,0040) Other Patient IDs (0010,1000) Ethnic Group (0010,2160) The FORUM MWL AE returns one of the following status codes in a C-FIND response: Service Status Further Meaning Error Code Description Refused No Such SOP class 0118 If a SOP Class is used, which was not negotiated. Unrecognized Operation 0211 Unrecognized value in Command Filed Failed Processing Failure 0110 Unexpected internal errors. Unable To Process C000 Unexpected internal errors. Cancel FE00 The original requester canceled this operation. Pending Pending FF00 All Optional Keys are supported in the same manner as Required Keys. Success Success 0000 Operation performed properly. Note: MWL entries (Scheduled Procedure Steps) are removed from the FORUM MWL AE as soon as the first SOP Instance with the Study Instance UID from the respective Requested Procedure is received. Furthermore MWL entries (Scheduled Procedure Steps) which are created via the "Automatic Order Functionality" or via the EMR Gateway are removed from the FORUM MWL AE at midnight. Document: DICOM Conformance Statement_FORUM_3.1.doc Page 18 of 25
19 4.3 Network Interfaces Physical Network Interface The physical network interface is not visible for the applications. The application uses the communication stack as offered by the Operating System Additional Protocols No additional protocols are supported. Document: DICOM Conformance Statement_FORUM_3.1.doc Page 19 of 25
20 4.4 Configuration AE Title/Presentation Address Mapping The translation from Application Entity Title to Presentation Address (TCP/IP address) is stored in the local database. Along with this mapping, the database stores those AE titles that are allowed to communicate with the FORUM AE and the FORUM MWL AE Local AE Titles The IP is not configurable. The IP is administrated by the Operating System. The AE titles for the FORUM AE and the FORUM MWL AE are configurable. The FORUM application can be configured to use only one AE for all transactions Remote AE Titles The mapping of external AE Titles to TCP/IP addresses and ports is configurable. The number of AE Titles is limited according to the available licenses, where it is distinguished between "Zeiss Devices" and "3 rd Party Devices" Logging Via the FORUM Service Tool it is possible to activate an extended DICOM logging. Document: DICOM Conformance Statement_FORUM_3.1.doc Page 20 of 25
21 5 Media Interchange Media Interchange is not scope of this document since Media Interchange is not supported by FORUM. Document: DICOM Conformance Statement_FORUM_3.1.doc Page 21 of 25
22 6.1 Accepted Character Sets 6 Support Of Character Sets Supported Specific Character Set Character Set Description Defined Term US-ASCII ISO ISO_IR 100 ISO ISO_IR 101 ISO ISO_IR 109 ISO ISO_IR 110 ISO ISO_IR 144 ISO ISO_IR 127 ISO ISO_IR 126 ISO ISO_IR 138 ISO ISO_IR 148 JIS_X0201 ISO_IR 13 TIS-620 ISO_IR 166 US-ASCII ISO 2022 IR 6 ISO ISO 2022 IR 100 ISO ISO 2022 IR 101 ISO ISO 2022 IR 109 ISO ISO 2022 IR 110 ISO ISO 2022 IR 144 ISO ISO 2022 IR 127 ISO ISO 2022 IR 126 ISO ISO 2022 IR 138 ISO ISO 2022 IR 148 JIS_X0201 ISO 2022 IR 13 TIS-620 ISO 2022 IR 166 JIS0208 ISO 2022 IR 87 JIS0212 ISO 2022 IR 159 cp949 ISO 2022 IR 149 UTF-8 ISO_IR 192 GB18030 GB Returned Character Sets DICOM files will be returned as they are. In case of a modified DICOM file it will be changed to UTF-8 and will be returned with this character set. No transcoding will be done with one exception: UTF-8 files will be returned as "ISO_IR 100" if the C-MOVE destination AE title contains the string "IR100". Document: DICOM Conformance Statement_FORUM_3.1.doc Page 22 of 25
23 7 Security The DICOM capabilities of the FORUM Application do not support any specific security measures. It is assumed that FORUM is used within a secured environment. It is assumed that a secured environment includes at a minimum: Firewall or router protections to ensure that only approved external hosts have network access to FORUM. Firewall or router protections to ensure that FORUM only has network access to approved external hosts and services. Any communication with external hosts and services outside the locally secured environment use appropriate secure network channels (e.g. such as a Virtual Private Network (VPN)) Other network security procedures such as automated intrusion detection may be appropriate in some environments. Additional security features may be established by the local security policy and are beyond the scope of this conformance statement. Document: DICOM Conformance Statement_FORUM_3.1.doc Page 23 of 25
24 8 Annexes 8.1 IOD Contents Created SOP Instance(s) FORUM creates new instances for manually imported images or PDF's. FORUM creates new instances for "Combined Reports" either automatically according to certain rules or manually Usage of Attributes from Received IODs FORUM uses the attribute in the following sequence: If Acquisition DateTime is not filled, then Acquisition Date and Time will be used as FORUM Acquisition DateTime. If Series Date and Time are not filled, then Study Date and Time will be used as FORUM Series Date and Time. If FORUM Series Date and Time are still not filled, then Content Date and Time will be used as FORUM Series Date and Time. If FORUM Acquisition DateTime is still not filled, then FORUM Series Date and Time will be used as FORUM Acquisition DateTime. If Image Laterality is not filled, then Laterality will be used as FORUM Image Laterality Attribute Mapping Coerced/Modified Files Patient Information The following terms are defined: Strict Attributes: - Patient ID - Patient's Name Last Name - Patient's Name First Name - Patient Birth Date Lax Attributes: - Issuer of Patient ID - Patient's Name Prefix - Patient's Name Suffix - Patient's Name Middle Name - Patient's Sex Matching Lax Attribute: - Lax attributes match if they are equal - Lax attributes match if they are empty on either side Corresponding Patient: - Corresponding patients have equal strict attributes - Corresponding patients have matching lax attributes Patient Information Attributes: - All Strict Attributes - All lax attributes - Ethnic Group, - Other Patient IDs Document: DICOM Conformance Statement_FORUM_3.1.doc Page 24 of 25
25 - Patient Comments Patient Information will be updated in the DICOM files: - If the user edits one of the patient information attributes - If an incoming DICOM File has an existing corresponding patient in FORUM and differences in the patient information attributes Missing UIDs Study Instance UID, Series Instance UID and SOP Instance UID will be auto generated if one of those is missing in an incoming DICOM file Patient Identification Conflict with Study Instance UID If an incoming DICOM file belongs to an existing study in FORUM and this existing study belongs to a different existing patient (different patient = not "corresponding patient", see ) then a new Study Instance UID will be auto generated. 8.2 Data Dictionary of Private Attributes FORUM does not create any private attributes. 8.3 Private Transfer Syntaxes No Private Transfer Syntaxes are supported. Document: DICOM Conformance Statement_FORUM_3.1.doc Page 25 of 25
DICOM Conformance Statement CBS Images and Worklist Version 2.01
DICOM Conformance Statement CBS Images and Worklist Version 2.01 Compulink Inc. 1100 Business Center Circle Thousand Oaks, CA 91320 Internal document number: 4226-000-000-001 rev 1 Creation Date: 20131101
Technical Publication. DICOM Conformance Statement. DICOM Proxy 2.0. Document Revision 3. October 20, 2010. 2010 Copyright Brainlab AG
Technical Publication DICOM Conformance Statement Document Revision 3 October 20, 2010 2010 Copyright Brainlab AG 1 Conformance Statement Overview This is a conformance statement for the Brainlab software
DICOM Conformance Statement
Date: 2010-10-7 DICOM Conformance Statement Stratus OCT Instrument and Review Software Version 7.0 Carl Zeiss Meditec, Inc. 5160 Hacienda Drive Dublin, CA 94568 USA www.meditec.zeiss.com Document: DICOM_Conformance_Statement_Stratus_7.0.doc
DICOM 3.0 Conformance Statement
Version 2.7 DICOM 3.0 Conformance Statement 1 1. Conformance Statement Overview SOP Classes Query and Retrieve Study Root Query/Retrieve Information Model Find Study Root Query/Retrieve Information Model
DICOM Conformance Statement
Date: 2014-12-18 DICOM Conformance Statement CIRRUS TM HD-OCT 5000 and 500 Instrument Version 7.5/8.0 Carl Zeiss Meditec, Inc. 5160 Hacienda Drive Dublin, CA 94568 USA www.meditec.zeiss.com Document: DICOM_Conformance_Statement_CIRRUS_7.5_8.0.Docx
Version 8 DICOM Conformance Statement. Version 3.04, September 2014
Version 8 DICOM Conformance Statement Version 3.04, September 2014 1 Conformance Statement Overview The application described in this Conformance Statement VEPRO EMR Manager PACS is a collection of processes
MiPACS Storage Server Conformance Statement Version 1.02.0.000
MiPACS Storage Server Conformance Statement Version 1.02.0.000 December 2009 Version 1.00.4.001 Conformance Statement Document Information Issued by: Medicor Imaging 1927 S. Tryon St., Suite 200 Charlotte,
Candelis, Inc. DICOM Conformance Statement. ImageGrid Storage Server
18821 Bardeen Ave. Irvine, CA 92612 Phone: 800.800.8600 Fax: 949.752.7317 Candelis, Inc. DICOM Conformance Statement ImageGrid Storage Server 0 INTRODUCTION ImageGrid Storage Server is a complete hardware/software,
Infinity Medical Image Server
Infinity Medical Image Server DICOM Server Conformance Statement Part Number: 900-420-002 Rev. 02 Document/Software Version: v1.3.0 Date: May 12, 2009 Copyright 2008-2009 Codonics, Inc. All Rights Reserved.
DOLPHIN DICOM IMAGING DICOM CONFORMANCE STATEMENT
DOLPHIN DICOM IMAGING DICOM CONFORMANCE STATEMENT NOTICE The software described in this document has been validated in accordance with the governing DICOM standard at the time of this document's release.
ONIS 2.0 DICOM CLIENT. DICOM 3 Conformance statement
ONIS 2.0 DICOM CLIENT DICOM 3 Conformance statement Table of Contents 0 Introduction...4 0.1 Scope...4 0.2 Standards...4 0.3 Intended Audience...4 0.4 Important Remarks...4 0.5 Definitions...4 1 Implementation
DICOM Conformance Statement. Version: 1.0
DICOM Conformance Statement Version: 1.0 Table of Contents 1 Preface...4 2 Overview...5 3 Introduction...5 3.1 Revision History...5 3.2 Audience...5 3.3 Remarks...6 3.4 Definitions, Terms and Abbreviations...6
Medflow Imaging DICOM Server
DICOM Server version 7.6.2 Revision 2.0 Copyright 2011 Medflow, Inc. Revision History Date June 14, 2011 August 30, 2011 Rev Description 7.6.1 First version. 7.6.2 Revised to reflect version 7.6.2 changes
HP Medical Archive Solutions DICOM Conformance Statement. January 2007 (Third Edition) Part Number T4434-96003
HP Medical Archive Solutions DICOM Conformance Statement January 2007 (Third Edition) Part Number Copyright 2007, 2007 Hewlett-Packard Development Company, L.P. Confidential computer software. Valid license
DICOM Conformance Statement Merge Eye Care PACS v. 4.0
DICOM Conformance Statement Merge Eye Care PACS v. 4.0 Merge Healthcare 900 Walnut Ridge Drive Hartland, WI 53029 USA 877.44.MERGE 2012 Merge Healthcare. The information contained herein is confidential
GE PACS Conformance Statement for DICOM v3.0
g GE Medical Systems Technical Publications IIS P/N 4361668 Revision 01 GE PACS Conformance Statement for DICOM v3.0 Copyright 1998 By General Electric Company g GE Medical Systems GE Medical Systems,
Dx MM DICOM 3.0 Conformance Statement
Dx MM DICOM 3.0 Conformance Statement NAME FONCTION VISA AUTHOR Alain Battistini Senior Engineer CONFORM TO ORIGINAL REVIEW Jean-François Hoh Senior Engineer CONFORM TO ORIGINAL APPROVAL Pascal Gilbert-Jeantet
DICOM Conformance Statement
Inter Medical Medizintechnik GmbH Daimlerstr. 34-36 D-32312 Lübbecke Tel. (0 57 41) 23 59-0 Fax (0 57 41) 23 59-10 E-Mail Contact @Intmed.de www.intmed.de Doc. Rev. 1 GammaXP DICOM Conformance Statement
g GE Medical Systems Advantage Cluster Storage / Archive System Release 1 v1.1.5.2 Conformance Statement Direction number: 2160628 Revision: 2
g GE Medical Systems Advantage Cluster Storage / Archive System Release 1 v1.1.5.2 Conformance Direction number: 2160628 Revision: 2 Date: September 20, 1996 Revision History Revision Date Author Reason
Hologic Physician s Viewer 7.0 DICOM Conformance Statement
Hologic Physician s Viewer 7.0 DICOM Conformance Statement MAN-02386 REV 001 Page 1 of 32 1. Conformance Statement Overview This document is the DICOM Conformance Statement for Hologic Physician s Viewer,
DICOM Conformance Statement
Document Number: 67923 Revision: B Date: 2009-01-15 Author: Sonia Yu DICOM Conformance Statement Cirrus HD-OCT Instrument and Review Software Version 4.0 Carl Zeiss Meditec, Inc. 5160 Hacienda Drive Dublin,
DICOM Conformance Statement. GDxPRO
Document Number: 2660021133317 Revision: D Date: 01/20/2010 Author: Vidya Kulkarni DICOM Conformance Statement GDxPRO GDxPRO Instrument Software Version 1.1 GDxPRO Review Software Version 1.0 Carl Zeiss
DICOM Conformance Statement For Diagnostic Review Workstation Software Version 5-x MAN-00546
DICOM Conformance Statement For Diagnostic Review Workstation Software Version 5-x MAN-00546 SecurView DX DICOM Conformance Statement Part Number MAN-00546 Revision 001 For Software version 5-X Service
CARESTREAM PACS Suite (Client, Server, and CD Direct) Version 11.0. DICOM Conformance Statement
CARESTREAM PACS Suite (Client, Server, and CD Direct) Version 11.0 DICOM Conformance Statement Publication # 8G8913 22 April 2009 Table of Contents Table of Contents...2 1 Introduction...5 1.1 Abbreviations
DigitizingStation. DICOM V3.0 Conformance Statement
DigitizingStation DICOM V3.0 Conformance Statement Copyright 2004 by image diagnost GmbH Page 1 of 17 Image Diagnost International GmbH Reproduction of this document without prior written approval is prohibited.
ClearCanvas ImageServer DICOM Conformance Statement
ClearCanvas ImageServer DICOM Conformance Statement Document Version: 2.2 Product Name(s): ClearCanvas ImageServer Release: 11.0 Date: January 12, 2014 ClearCanvas Inc. 1920 439 University Ave. Toronto,
DICOM 3.0 Conformance Statement
Picture Archiving and Communication System DICOM 3.0 Conformance Statement CODE NO.1245 Manufacturer: 1 Sakura-machi, Hino-shi Tokyo 191-8511, Japan Important Notes Konica Minolta Medical & Graphic, Inc.
AquariusNET DICOM Conformance Statement. DICOM Conformance Statement. AquariusNET 4.4. Rev. 12.03.2015 B
DICOM Conformance AquariusNET 4.4 Rev. 12.03.2015 B 1 1 Conformance Overview The AquariusNET server is a 2D/ 3D enterprise thin client-server solution. This central 3D processing server streams real-time
DICOM Conformance Statement
This printed copy is not subject to any change control XXXXXXXXXXXXxxXXXXXXXXXXXXXXXXXXX MR 2013-03-14 Siemens AG 2013 All rights reserved Siemens AG Healthcare Sector, Henkestr. 127, D-91050 Erlangen,
Technical Publications
Technical Publications Direction DOC0708777 Revision 3 Centricity Enterprise Archive V4.0 DICOM CONFORMANCE STATEMENT Copyright 2011 by General Electric Company 1 CONFORMANCE STATEMENT OVERVIEW The Enterprise
Xeleris 2.0 Conformance Statement for DICOM V3.0
g GE Healthcare Technical Publications DIRECTION 5168972-100 Revision 2 Xeleris 2.0 Conformance Statement for DICOM V3.0 Document Structure Information: The DICOM Print Services for Xeleris are defined
DICOM Conformance Statement
DICOM Conformance Statement CARESTREAM Vue PACS v11.3.2.4 Lesion Management and Bookmark Applications Part # 6K4188 2012-07-02 PAGE 1 of 142 Table of Contents 1 INTRODUCTION... 4 1.1 ABBREVIATIONS AND
DICOM Conformance Statement
Document Number: CG-S2F-MS-039 Revision: 1.0 Date: 2011-05-31 DICOM Conformance Statement FORUM LINK net Version 1.0 Carl Zeiss Meditec AG Göschwitzer Strasse 51-52 07745 Jena Germany www.meditec.zeiss.com
AGFA MEDICAL IMAGING DICOM Conformance Statement
MED/HB/000405 Page 1 of 19 AGFA MEDICAL IMAGING DICOM Conformance Statement ADC PRID - Software Version 1.1.09 SCU Worklist Management Page 2 of 19 MED/HB/000405 18 September, 2001 Agfa Medical Imaging
Technical Publications
g GE Healthcare Technical Publications Direction 5357330-1EN (DOC1498472) Revision 3 Xeleris 3.0 and Xeleris 3.1 DICOM CONFORMANCE STATEMENT Copyright 2014 by General Electric Co. Do not duplicate g GE
DICOM Conformance Statement
DICOM Conformance Statement Page 1 of 14 1 Introduction... 3 2 Quick Summary... 4 3 Storing Studies... 6 3.1 Ideal Store scenario...6 3.2 Non-ideal scenarios...6 3.2.1 Accession number...7 3.2.2 No indication
Technical Publications
g GE Healthcare Technical Publications Direction 1007228 Revision 0 Navigation Plus System Version 5.2 For DICOM Copyright 2004 by General Electric Co. Do not duplicate THIS PAGE LEFT INTENTIONALLY BLANK
Technical Publications
g GE Healthcare Technical Publications Direction DOC0855383 Revision 2 Centricity TM RIS-i MedoraCom 4.2 Plus Copyright 2011 by General Electric Co. Do not duplicate REVISION HISTORY REV DATE REASON FOR
Philips Medical Systems DICOM Conformance Statement
Philips Medical Systems DICOM Conformance Statement DSI Pro 1.0 Document Number 4522 164 07221 20 February 2001 Copyright Philips Medical Systems Nederland B.V. 2001 Page ii DICOM Conformance Statement
DICOM Conformance Statement
Revision: 1.0 Date: Sep, 15th 2015 DICOM Conformance Statement VISUCAM 224 and VISUCAM 524 Version 6.0 Carl Zeiss Meditec AG Goeschwitzer Strasse 51-52 07745 Jena Germany www.zeiss.com/med Document: DICOM_Conformance_Statement_VISUCAM_224_524_6.0.doc
DICOM CONFORMANCE STATEMENT FOR ZIOSTATION 2.0
DICOM CONFORMANCE STATEMENT FOR ZIOSTATION 2.0 DICOM Conformance Statement, Ziostation 2.0 1 Copyright 2006-2011, Ziosoft, Inc. 0 DICOM CONFORMANCE STATEMENT OVERVIEW 4 0.1 APPLICATION DATA FLOW... 5 0.2
AGFA HEALTHCARE DICOM Conformance Statement
Page 1 of 79 AGFA HEALTHCARE DICOM Conformance Statement IMPAX 6.5 Solution Document No. 001310 Revision: 1.1 When printed, this is NOT a controlled document. Page 2 of 79 Document Information Service-related
Digital Imaging and Communications in Medicine (DICOM) Supplement 119: Frame Level Retrieve SOP Classes
Page i 5 Digital Imaging and Communications in Medicine (DICOM) Supplement 119: Frame Level Retrieve SOP Classes 10 15 20 Prepared by: Working Group 6 DICOM Standards Committee, Working Group 6 1300 N.
DICOM CONFORMANCE STATEMENT STORAGE SCU, Q/R SCP, PRINT SCU & STORAGE COMMITMENT SCU FOR TOSHIBA SUPERCONDUCTING MRI SYSTEMS (MIIMR0001EAB)
DICOM CONFORMANCE STATEMENT STORAGE SCU, Q/R SCP, PRINT SCU & STORAGE COMMITMENT SCU FOR TOSHIBA SUPERCONDUCTING MRI SYSTEMS FLEXART-series/VISART-series/ EXCELART-series/SECOND CONSOLE (MIIMR0001EAB)
DICOM 3.0 CONFORMANCE STATEMENT
DICOM 3.0 CONFORMANCE STATEMENT JETStream Workspace Version 1.0 9705-0067, Rev A October, 004 Copyright Statement Philips Medical Systems has taken care to ensure the accuracy of this document. However,
MammoView 1.5. DICOM Conformance Statement 1.0. medigration GmbH All rights reserved
All rights reserved medigration GmbH, Schuhstr. 30, 91052 Erlangen Tel: +49 (0) 9131/69087-40, Fax: +49 (0) 9131/69087-50 www.medigration.de Copyright. All rights reserved. Table of Contens 1 Introduction...
DICOM CONFORMANCE STATEMENT FOR ZIOCUBE 1.0
DICOM CONFORMANCE STATEMENT FOR ZIOCUBE 1.0 DICOM Conformance Statement, ZioCube 1.0 1 Ziosoft, Inc. 0 DICOM CONFORMANCE STATEMENT OVERVIEW 4 0.1 APPLICATION DATA FLOW...5 0.2 FUNCTIONAL DEFINITION OF
GENIE Acquisition R3.1 Conformance Statement for DICOM v3.0
Technical Publications Direction 2358711-100 Revision 0 GENIE Acquisition R3.1 Conformance Statement for DICOM v3.0 This document applies to the following systems: Millennium MPR/MPS, MG and MC Optima
Varian System Server. DICOM Conformance Statement
Varian System Server DICOM Conformance Statement P/N VA8005D3CS July 2007 General Information Abstract This document provides information about the DICOM Conformance of the Varian System Server product
Technical Publications
g GE Healthcare Technical Publications Direction DOC1486636 Revision 2 Optima CT Dicom Conformance Statement for DICOM * Refer to section 1 for a list of products in the Optima CT families to which this
How To Connect Ifa Dicom To An Ophthalmology System
High Level Health IT (HIT) Standards for Ophthalmologists worldwide DICOM : the comprehensive ifa Solutions (Digital Imaging and Communication in Medicine) 1 1 2 3 4 5 6 7 8 Overview Introduction and DICOM
Centricity TM RISi DICOM Conformance Statement
Technical Publications Direction No 2118780 Release 3.12 and 4.2 - General Electric Company - All rights reserved. Table of Contents 1 Introduction 4 1.1 Overview 4 1.2 OVERALL DICOM CONFORMANCE STATEMENT
HDI 4000 Ultrasound System
HDI 4000 Ultrasound System DICOM Conformance Statement Rev 1.0 Philips Medical Systems HDI 4000 DICOM Conformance Rev. 1.0 12/17/2001 Philips Medical Systems Table of Contents 0. Introduction...3 0.1 DICOM
Table of Contents. Version History
Table of Contents 1. Introduction... 1 2. Implementation Model... 1 2.1 Application Data Flow Diagram... 1 2.2 Functional Definition of AE s... 2 2.3 Sequencing of Real World Activities... 2 3. AE Specifications...
DICOM Conformance Statement
Philips Medical Systems DICOM Conformance Statement Integris Systems with High Speed DICOM Image Interface MCV 2974 MCV 3621 Cardiac DICOM XA MCV 3761 Vascular DICOM XA MCV 3031 DICOM RIS Interface and
DICOM. Conformance Statement. Envisor Software Version C.0
DICOM Conformance Statement Envisor Software Version C.0 COPYRIGHT KONINKLIJKE PHILIPS ELECTRONICS N.V. 2005. ALL RIGHTS RESERVED. NO PART OF THIS PUBLICATION MAY BE REPRODUCED OR TRANSMITTED IN ANY FORM
Technical Publications
GE Healthcare Technical Publications Direction 5720690-1EN Revision 1 AW Server 3.2 CONFORMANCE STATEMENT for DICOM do not duplicate Copyright 2015 By General Electric Co. THIS PAGE LEFT INTENTIONALLY
Printlink5-ID_IV. DICOM 3.0 Conformance Statement PRINT MANAGEMENT SYSTEM CODE NO. 1118. Manufacturer:
PRINT MANAGEMENT SYSTEM Printlink5-ID_IV DICOM 3.0 Conformance Statement CODE NO. 1118 Manufacturer: 2970 Ishikawa-machi, Hachioji-shi Tokyo 192-8505 Japan DICOM Conformance Statement Important Notes Konica
Centricity Enterprise Web 3.0 DICOM Conformance Memo DOC0094970
DOC0094970 CONTENTS 1 Introduction... 3 1.1 Scope and Purpose... 3 1.2 Intended Audience... 3 1.3 Scope and Field of Application... 3 1.4 References... 4 1.5 Definitions... 4 1.6 Symbols and Abbreviations...
OpenEMR: Achieving DICOM Interoperability using Mirth
OpenEMR: Achieving DICOM Interoperability using Mirth A ViSolve, Inc. Technical Guide TABLE OF CONTENTS Table of Contents 1. Objective... 3 2. DICOM Images... 3 3. DICOM Image Viewers... 4 4. Sending and
DICOM CONFORMANCE STATEMENT
Technical Publications Revision 3 Reporting Tool 2.7 DICOM CONFORMANCE STATEMENT Copyright 2005-2010 by General Electric Co. Do not duplicate Page 1 of 34 REVISION HISTORY Revision Date Reason for Change
DICOM Conformance Statement. DICOMscope 3.5. Software developed by: M. Eichelberg 1, K. Kleber 2, J. Riesmeier 1, A. Schröter 2, A.
DICOM Conformance Statement DICOMscope 3.5 Software developed by: M. Eichelberg 1, K. Kleber 2, J. Riesmeier 1, A. Schröter 2, A. Thiel 1 Organizational support: J. Holstein 2, H. Oosterwijk 3 1) Kuratorium
Digital Imaging and Communications in Medicine (DICOM) Part 4: Service Class Specifications
Digital Imaging and Communications in Medicine (DICOM) Part 4: Service Class Specifications Published by National Electrical Manufacturers Association 1300 N. 17th Street Rosslyn, Virginia 22209 USA Copyright
How To Write A Dicom Dicoma Dicomm Test Article
Technical Publications Direction 2183325 100 Revision 0, March 27, 1997 GENIE Processing & Review System, R1.7 DICOM 3.0 Conformance Statement Document Structure Information: The DICOM Print Services for
ConnectPro for CT/i 4.0 Systems Conformance Statement for DICOM v3.0
Technical Publications Direction 2210573-100 Revision 0 ConnectPro for CT/i 4.0 Systems Conformance Statement for DICOM v3.0 Copyright 1998 by General Electric Company GE Medical Systems: Telex 3797371
DICOM Conformance Statement
DICOM Conformance Statement IntelliSpace Breast R3.1 Multi-modality features powered by IntelliSpace Clinical Applications R9.1V7 Koninklijke Philips Electronics N.V. 2013 All rights are reserved. Issued
JiveX Enterprise PACS Solutions. JiveX HL7 Gateway Conformance Statement - HL7. Version: 4.7.1 As of 2015-05-20
JiveX Enterprise PACS Solutions JiveX HL7 Gateway Conformance Statement - HL7 Version: 4.7.1 As of 2015-05-20 VISUS Technology Transfer GmbH Universitätsstr. 136 D-44799 Bochum Germany Phone: +49 (0) 234
Extracting, Storing And Viewing The Data From Dicom Files
Extracting, Storing And Viewing The Data From Dicom Files L. Stanescu, D.D Burdescu, A. Ion, A. Caldare, E. Georgescu University of Kraiova, Romania Faculty of Control Computers and Electronics www.software.ucv.ro/en.
DICOM Conformance Statement
DICO Conformance Statement IDS5 11.4; IDS7, Sectra Healthcare Server, WISE, ImageServer/s, ImageServer/xd Version 12.2.3, June 2010 About this document Sectra Imtec AB, Sweden, 2010 Disclaimer Sectra Imtec
PARCA Certified PACS Interface Analyst (CPIA) Requirements
PARCA Certified PACS Interface Analyst (CPIA) Requirements Copyright notice: Copyright 2005 PACS Administrators in Radiology Certification Association (PARCA). All rights reserved. All rights reserved.
DICOM Conformance Statement
Carestream DryView 5700 Laser Imager DICOM Conformance Statement June 15, 2011 Document # 8J5397 PAGE 1 of 40 This publication is protected by Federal Copyright law, with all rights reserved. Note: Publication
DICOM CONFORMANCE STATEMENT
Technical Publications Revision 2 Reporting Tool DICOM CONFORMANCE STATEMENT Copyright 2005-2009 by General Electric Co. Do not duplicate Page 1 of 31 REVISION HISTORY Revision Date Reason for Change A
DICOM Conformance Statement. Veradius Unity
DICOM Conformance Statement Veradius Unity Koninklijke Philips Electronics N.V. 2014 All rights are reserved. Document Number: ICAP-PF.0008674 Issued by: Philips Healthcare Philips Nederlands Best P.O.
IHE Radiology Technical Framework Supplement. Web-based Image Capture (WIC) Draft for Public Comment
Integrating the Healthcare Enterprise 5 IHE Radiology Technical Framework Supplement 10 Web-based Image Capture (WIC) 15 Draft for Public Comment 20 Date: February 19, 2015 Author: IHE Radiology Technical
HL7 Conformance Statement
HL7 Conformance Statement Release VA20B (2014-03-28) ITH icoserve technology for healthcare GmbH Innrain 98, 6020 Innsbruck, Austria +43 512 89059-0 www.ith-icoserve.com Any printout or copy of this document
MedDream SendToPACS USER S MANUAL (version 1.15)
MedDream SendToPACS USER S MANUAL (version 1.15) 2014, Softneta UAB, Kaunas All rights reserved in the event of granting of patents or registration as a utility patent. All names of companies and products
Digital Imaging and Communications in Medicine (DICOM) Part 1: Introduction and Overview
Digital Imaging and Communications in Medicine (DICOM) Part 1: Introduction and Overview Published by National Electrical Manufacturers Association 1300 N. 17th Street Rosslyn, Virginia 22209 USA Copyright
NEMA Standards Publication PS 3 Supplement 41. Digital Imaging and Communications in Medicine (DICOM) Digital Signatures
NEMA Standards Publication PS 3 Supplement 1 Digital Imaging and Communications in Medicine (DICOM) Digital Signatures Status: Final Text Sep 001 Prepared by DICOM Standards Committee, Working Group 1
DICOM Correction Proposal
DICOM Correction Proposal STATUS Assigned Date of Last Update 20165/01/08 Person Assigned Submitter Name Harry Solomon Harry Solomon Submission Date 2015/09/21 Correction Number CP-1550 Log Summary: Refactor
DICOM Conformance Statement Print Management Service Class FUJI NETWORK PRINT SERVER FN-PS551
DICOM Conformance Statement Print Management Service Class FUJI NETWORK PRINT SERVER FN-PS55 October 006 5th Edition Copyright FUJIFILM Corporation, Japan 897N00094A (006-03-40) 006.0 Revision History
Tools for DICOM Implementation
DICOM INTERNATIONAL CONFERENCE & SEMINAR Oct 9-11, 2010 Rio de Janeiro, Brazil Tools for DICOM Implementation David Clunie CoreLab Partners, Inc. Outline Tools for DICOM implementation Toolkits and sample/reference
CT RADIATION DOSE REPORT FROM DICOM. Frank Dong, PhD, DABR Diagnostic Physicist Imaging Institute Cleveland Clinic Foundation Cleveland, OH
CT RADIATION DOSE REPORT FROM DICOM Frank Dong, PhD, DABR Diagnostic Physicist Imaging Institute Cleveland Clinic Foundation Cleveland, OH CT Patient comes out... Patient goes in... Big Black Box Radiology
DICOM Digital Imaging and Communications in Medicine
DICOM Digital Imaging and Communications in Medicine INF5081 / Vårsemester 2007 Wolfgang Leister Reidar Øksnevad Pedro Cruz G. April 2007 PACS PACS = Picture Archive and Communication System Archive for
DV5950 DICOM Conformance Statement. Product: DryView 5950 Laser Imaging System
DV5950 DICOM Conformance Statement Product: DryView 5950 Laser Imaging System Document # 9J5331 Author: Shobha Murthy This publication is protected by Federal Copyright law, with all rights reserved. Note:
Digital Imaging and Communications in Medicine (DICOM) Part 5: Data Structures and Encoding
Digital Imaging and Communications in Medicine (DICOM) Part 5: Data Structures and Encoding Published by National Electrical Manufacturers Association 1300 N. 17th Street Rosslyn, Virginia 22209 USA Copyright
Cenova Image Analytics Server DICOM Conformance Statement
breast imaging solutions DICOM Standard Interface Cenova Image Analytics Server DICOM Conformance Statement Software Version 2.3 MAN-03715 rev 001 Cenova Image Analytics Server 2.3 DICOM Conformance Statement
Kodak PACS Link Medical Image Manager 200 Kodak PACS Link 25 Print Server as Print Server (SCP) Software Version 6.1.1
Kodak PACS Link Medical Image Manager 200 Kodak PACS Link 25 Print Server as Print Server (SCP) Software Version 6.1.1 DICOM Conformance Statement June 30, 2005 Document # 3E5252, Rev. J Revision History
QueryRetrieveSCU Queries image archives and controls remote retrieve of images to specified destination. (= SCU of Query/Retrieve SOP classes).
8VLQJ-',&20,QWURGXFWLRQ JDICOM is a summary of DICOM applications which are very useful for troubleshooting. Our best appreciation to Mr. Gunter Zeilinger which created this great tool. If you have any
DICOM: Definitions and Testing
DICOM: Definitions and Testing Bruce Curran University of Michigan (AAPM Representative, DICOM WG-7 Walter R. Bosch Advanced Technology Consortium Washington University of St. Louis Acknowledgements Sam
NETWORK MANAGEMENT FOR PICTURE ARCHIVING AND COMMUNICATION SYSTEMS
NETWORK MANAGEMENT FOR PICTURE ARCHIVING AND COMMUNICATION SYSTEMS by Edlic Nga-Lik Yiu B.A.Sc., Simon Fraser University, 2000 Edwood Nga-Wood Yiu B.A.Sc., Simon Fraser University, 2000 PROJECT SUBMITTED
DICOM Conformance Statement. CT Tomoscan M/EG/EG Compact Release 4.0. Document Number 4522 220 59691 30 March 2000
Philips Medical Systems DICOM Conformance Statement CT Tomoscan M/EG/EG Compact Release 4.0 Document Number 4522 220 59691 30 March 2000 Copyright Philips Medical Systems Nederland B.V. 2000 Philips Medical
Patient Database and PACS Communication Manual
Patient Database and PACS Communication Manual January 28, 2016 Software platform v2.0 R4870 MEDVISO AB http://www.medviso.com Griffelvägen 3 SE-224 67 Lund Sweden Tel: +46-76-183 6442 ii Contents 1 Conventions
Implementation of DICOM Modality Worklist at Patient Registration Systems in Radiology Unit
EPJ Web of Conferences 68, 000028 (2014) DOI: 10.1051/ epjconf/ 20146800028 C Owned by the authors, published by EDP Sciences, 2014 Implementation of DICOM Modality Worklist at Patient Registration Systems
SIGNA PROFILE CONFORMANCE STATEMENT
Technical Publications Direction 2174981 Revision 1.0 SIGNA PROFILE CONFORMANCE STATEMENT for DICOM v3.0 (ID/Net v3.0) sm - Service Manual do not duplicate Copyright 1996 by General Electric Company blank
FUJIFILM MEDICAL SYSTEMS
FUJIFILM MEDICAL SYSTEMS Product Data Synapse Release Version 4.2 Server Modules and Interfaces Application Synapse server is a collection of software modules built on Microsoft Windows Server, which together